Inaccurate measurements in OMERO Insight

Postby hgoff » Wed Apr 16, 2014 1:26 pm

I measured a mouse tibia which according to the measurement tool was
55.1um, however I imagine that it would be in the order of mm (difficult
to say exactly how much without going back to check which mouse / age it
was, probably 10-15mm). Can you please advise if I have missed anything or wrong in process?
Thank you in advance
hgoff
 
Posts: 11
Joined: Wed Mar 19, 2014 10:58 am

Re: Inaccurate measurements in OMERO Insight

Postby jburel » Thu Apr 17, 2014 8:16 am

Thanks for reporting the problem.
It could be that we did not display correctly the value in the client or misread some information while reading the original image file. We will need the file to track down the issue.
Could you upload the file via our QA system http://qa.openmicroscopy.org.uk/qa/upload/
Which version of OMERO are you using?

Thanks
Jmarie
User avatar
jburel
Team Member
 
Posts: 348
Joined: Thu May 21, 2009 6:38 pm
Location: dundee

Re: Inaccurate measurements in OMERO Insight

Postby hgoff » Thu Apr 17, 2014 12:50 pm

Hi Jmarie, I am using OMERO 5. I am also double checking to make sure the image metadata is in the correct scale. once I confirm the settings on the machine are accurate I will upload the file to the QA system.

Thank you again for your help
hgoff
 
Posts: 11
Joined: Wed Mar 19, 2014 10:58 am

Re: Inaccurate measurements in OMERO Insight

Postby hgoff » Wed May 07, 2014 1:35 pm

Hi Jmarie, I have uploaded 2 DCM files to QA link. ID 9219. measurements are off in both in about a 1000. i.e. - 0.008mm instead of 8mm. additionally, this is xray image, why does it appear all black?

any help will be highly appreciated.
Thank you
hgoff
 
Posts: 11
Joined: Wed Mar 19, 2014 10:58 am

Re: Inaccurate measurements in OMERO Insight

Postby jburel » Thu May 08, 2014 2:09 pm

Thanks for uploading the file.
We will look at it asap.

Cheers
jmarie
User avatar
jburel
Team Member
 
Posts: 348
Joined: Thu May 21, 2009 6:38 pm
Location: dundee

Re: Inaccurate measurements in OMERO Insight

Postby hgoff » Fri May 16, 2014 9:23 am

Hi Jmarie,

I was wondering if you have managed to have a look at the dcm file?

Please let me know if additional information is required.
Thanks
Hilla
hgoff
 
Posts: 11
Joined: Wed Mar 19, 2014 10:58 am

Re: Inaccurate measurements in OMERO Insight

Postby jburel » Mon May 19, 2014 8:27 am

Hi Hilla

I did not have a chance yet, I will do that this week

Thanks

Jmarie
User avatar
jburel
Team Member
 
Posts: 348
Joined: Thu May 21, 2009 6:38 pm
Location: dundee

Re: Inaccurate measurements in OMERO Insight

Postby jburel » Thu May 22, 2014 12:52 pm

Hi Hilla

I checked your images. In both cases, the values stored for the pixels size X and size Y are 0.049 microns. This will explain the measurement you see when you draw lines in the Measurement tools.
There is probably an error while reading the metadata.
I checked using another tool and the distance should be in mm so we are not correctly reading the metadata, ticket opened see https://trac.openmicroscopy.org.uk/ome/ticket/12307

Regards
Jmarie
User avatar
jburel
Team Member
 
Posts: 348
Joined: Thu May 21, 2009 6:38 pm
Location: dundee

Re: Inaccurate measurements in OMERO Insight

Postby wmoore » Thu May 22, 2014 12:56 pm

If we're out by a factor of 1000 (as suggested above) are the real pixel sizes 0.049 mm (instead of microns)?
Is that metadata in the file anywhere?
If so, I guess it's just a Bio-Formats issue to convert to microns.
User avatar
wmoore
Team Member
 
Posts: 674
Joined: Mon May 18, 2009 12:46 pm

Re: Inaccurate measurements in OMERO Insight

Postby hgoff » Thu May 29, 2014 9:04 am

Thank you for looking at the files, great help!
I hope it will be planned for resolution soon.

Do you know why the image is completely black and not an x-ray?
if I export the DCM to TIFF it will appear as x-ray, and on the machine itself it is X-ray as well.

Thank you again for any feedback/input.
Hilla
hgoff
 
Posts: 11
Joined: Wed Mar 19, 2014 10:58 am

Next